Thanks Gaurav and Santam!
I think the major issue with using X Knife with multiple isocentres could perhaps be widening of the penumbra which does not seem to be a major issue with Radiosurgery.
Inherently, I am veering towards the idea of a heterogeneity of dose within a tightly confined volume but as Santam rightly points out, we might as well be hitting in the dark. But then, keeping the hot spots well in the target is not harming anyone; only we don
Now, is there any objective evidence that mutliple isocentres increase the conformality say in a given spherical volume without spilling off the excess dose? I can foresee a simple experiment to set up a spherical phantom with dosimetry to the be done at the edges and say to measure the distance between the prescription isodose and 50% as cut off. I could not find any objective published reference to my assertions.
One more issue. Planning multiple isocentres would probably increase the integral dose; not an issue in palliation but definitely for the benign cases. Has the use of mMlcs solved this problem? Or perhaps the use of newer TPS? In that case how do you go about it to reduce the integral doses in case multiple isocentres are required?